I recently purchased the SC-300 and forgot to screw on the pH storage bottle and probe was left to dry for a couple days. Is my pH probe ruined? If not, what should I do?

Place the electrode back in the storage solution as soon as possible. It may come back to life. Soak the probe in the storage solution (but make sure this solution is clean and clear) for a few days and then try and calibrate it and see what happens. If you do not have fresh storage solution you can soak it in the pH 4.01 reference solution until you are able to get fresh pH electrode storage solution.

My local store is suggesting I use a TA Titrant of 0,1 N or 0,2N (sodium hydroxide) as a replacement of the TA Titrant that came with my Vinmetrica SC-300 kit. Will I get the same results?

The 0.133N NaOH (TA Titrant) that we make is specifically designed to be used with our kit when testing your TA levels. You can use the 0.1N NaOH as a replacement for the TA Titrant however there will need to be adjustments made to your final calculations. You cannot use the simple equation listed at the end of the procedure (TA = 2 * V). You will need to use the longer equation listed above it and replace the 0.133N with the concentration of the NaOH solution that you use. This may blur your endpoint slightly so do take care when performing the titration and make sure to not overshoot the endpoint.

I get a “Good CAL” with 4.0 & 7.0 buffers and get stable readings on wine samples. However, I can never get a stable pH reading on distilled water. Is it possible to test the pH of distilled water?

A better way to test the validity of pH calibration is to measure the pH of a saturated solution of cream of tartar.

  1. Get pure cream of tartar (grocery store stuff is fine, provided it’s pure), or reagent grade potassium hydrogen tartrate, also known as potassium acid tartrate or potassium bitartrate. Call it KHT for short.
  2. Place about 1/8 teaspoon of KHT in 20 mL of distilled water. Mix well for about 30 seconds. You want to be sure the solution is saturated, i.e., everything that can dissolve, has dissolved. There should be some undissolved solid left.
  3. Decant or filter the solution off the solids if possible.
  4. This solution has a standard pH of 3.56 at 25 degrees C (78 degrees F). It should be within 0.02 pH of this value at temperatures from 20 to 30 degrees Celsius. We usually are OK with a value between 3.50 and 3.60. Discard after 24 hours

It is possible to measure the pH of distilled water, but it’s tricky. In principle pure water has a pH of 7.00 at normal temperatures, but as atmospheric CO2 dissolves in the water the pH will drop slightly due to the formation of carbonic acid. In addition, most distilled water from the store, (which by the way is usually de-ionized, rather than distilled), may have very minor amounts of acidity or alkalinity that takes the pH away from 7.00. So even if you degas and boil the water, you may find the pH still not hitting 7.00 and in any case it will start changing immediately on exposure to air.

However, the pH of your distilled water is unimportant for almost anything you want to do with it, because the water has negligible buffering ability. If you were to take 25 mL of it and do a TA titration for example, you will see that less than 1 drop of TA Titrant takes it to way above pH 9, so in essence it has no detectable acidity.

When testing TA, I typically get a lower pH reading after adding 15 ml distilled water to the 5 ml wine sample (i.e. wine is 3.49 pH and wine with 15 ml water is 3.42). This looks like my distilled water may be slightly acidic. Does this skew my final TA results?

The reason that the pH of wine drops on adding distilled water has nothing to do with the water or, up to a point, with the degree of dilution. Rather it has to do with the nature of the pH electrode and how it measures pH. pH is generally considered to be -log[H+], i.e., the negative logarithm, base 10, of the hydrogen ion concentration, but formally it is defined as -log(hydrogen ion activity). Activity a and concentration c are related but not identical: a = g x c where g is the “activity coefficient”. It is this coefficient that is changing as you dilute your wine, becoming closer to 1 with dilution. The pH electrode responds to hydrogen ion activity, rather than its concentration per se.

Also, since wine is a fairly strongly buffered system (primarily by the tartaric acid/tartrate ions, and other acids present) you would have to dilute it over a thousand fold to change the hydrogen ion concentration.

To check your pH calibration, suspend about 1/8 tsp (0.5 g) of pure cream of tartar (aka potassium hydrogen tartrate) in 20 mL of water, stir to dissolve for 30 sec. Some crystals will remain. This saturated solution of cream of tartar has a standard pH at 25C of 3.56. I am good with 3.50 to 3.60. If not, try recalibrating. Make this fresh each day.

 

I want to do a TA test, but the main acid is not Tartaric Acid. Can I report TA as Lactic Acid in beer or malic acid in cider ?

Yes! The SC-200 or SC-300 can be used to test beer, cider or any other beverages.  The pH is independent of the beverage or whatever solution it’s measuring.
The TA is also; it’s just a matter of how you calculate and report the final answer.   Wine TA is usually reported in terms of tartaric acid, but the Vinmetrica TA titration actually measures acidity as equivalents per liter (eq/L), converting it to g/L of tartaric acid by the equation

                                                                  TA (g/L tartaric) = 2 * V
as in the manual.  The equation assumes the acid is mostly tartaric (true for wine) and incorporates tartaric acid’s equivalent weight**  of 75 g/eq into its factor of 2.
For measuring the TA levels in a sour beer you would use the equivalent weight for lactic acid (90 g/eq) since that is the prevalent acid.  You can adjust  the result of using the above equation to report lactic acid instead of tartaric as follows:
                                                           g/L lactic = 90/75 * g/L tartaric.

For TA levels in cider, use the equivalent weight  67  for malic acid since that is the prevalent acid.

**Equivalent weight = molecular weight divided by the number of acid groups. Tartaric and malic acids both have two acid groups, so their equivalent weights are one-half of their molecular weights.  Lactic acid has only one acid group, so its equivalent weight and molecular weight are the same.